Hamilton HEPA Hood recommended?

Hi there,

as I’m having trouble with contaminated samples in our Hamilton Vantage from time to time, I’m considering purchasing the HEPA filtered hood as an additional safety net. Does anyone here have it (or the equivalent on a STAR) and can give me a recommandation on what to look out for?

I’m also looking at additional tips for keeping the inside of the robot clean. E.g. DNA Exitus Plus? We are already having UV raditation of 10 minutes each morning before we start working, but sometimes still seem to have Amplicon contamination in our negaitve controls.

Any advice appreciated,

best
Dominik

Hamilton’s CAP works well, treat it as a sterile positive pressure source rather than a BSC. As the air is turbulent, you should give it a bit of time to purge the dirty air at the start of methods. I vaguely remember being told not to run it all of the time.

1 Like

In my experience, HEPA-filtered hoods are more useful for applications like cell culture where maintaining a sterile environment is important.

For amplicon contamination, I’d focus on cleaning the deck, carriers and waste tray thoroughly with detergents like DNA-ExitusPlus and then rinse with water, etc.

If the instrument is used regularly, I’d also recommend daily cleaning of the waste plate and waste bag and weekly cleaning of the pipetting channels according to Hamilton’s guidelines.

I’d also look into whether droplets or splashing during aspiration, dispensing or tip disposal could be contributing to the contamination.

In our PCR lab, we UV the STARs for 30min daily. However, UV alone may not be sufficient, especially in areas that are shaded from direct exposure.

Personally, I’d investigate these potential contamination sources before investing in a HEPA hood, as filtration won’t necessarily address residual DNA contamination on the instrument surfaces.

2 Likes

UV is handy, but all those little shadowed spots are basically untouched.

Im not sure if a HEPA fan would help much since I expect this would mostly prevent external dna from getting inside the enclosure. There’s no guarantee of laminar flow, and increasing air flow in the chamber can actually increase dispersion of DNA.

How likely is it that you are seeing internal cross contamination vs external? Internal is more likely to derive from mixing, channel movement paths, sample plate movements, that kind of thing.

Another possible contamination vector would be if you have ever had an over aspiration event on any of the pipetting channels on either the individual pipettes or on the multichannel heads. It’s rare but it is a possibility.

An interesting test would be to stage a run with dummy samples and see if anything makes it in there.

1 Like

We are certain that it is not external DNA, since we now set up a rigid qPCR testing scheme to detect any amplicon produced in our lab. Since these are Illumina librarys I can check wether P5/P7 adapters are present in the contaminants. We ordered the hood anyways as there’s a higher concentration outside of the robot as our PCRs run independent of the Vantage. I only found one spot inside of the Vantage (one HHS unit where we has a spill once) that was contaminated. From our testing DNAExitus works wonders, after cleaning the benches and the robot with it we could not get any positive tests anymore. So I’m closing in on solving this but think that the hood will help to keep the “outside-air” outside :slight_smile:

1 Like