Blowout on tip pickup

Hi,

We are starting to use the Grenova washer to wash our Hamilton tips. And while the washing seems to work, we end up having a bit of water volume in the tip.
Is there a way with Venus (classical or firmware commands) to aspirate air before tip pickup and do a blowout once the tips are picked up to removed any liquid left in it ?

Thanks,

3 Likes

If I recall correctly, to do this one needs enable a special tip pickup property on the tip labware.

image

Then you would pick up your tips and then call a dispense blowout step.

2 Likes

Thank you,

I will try this today. Do you know if there’s a way to choose the volume ?
And where did you find that documentation ?

The labware properties (screenshot) are in the Programmers Manual. As for the knowledge of how this works, I am not sure if this info is documented anywhere particular but I asked Hamilton a similar question like 7 years ago when I was battling residual volume in tips from tip reuse.

Would love for a Hamilton rep to correct me if I am wrong on how this works

2 Likes

Thank you so much,

I forgot that part of the programmer’s manual and it is indeed very useful.

1 Like

Also u should reach out to Grenova. As they can update the program for u as we do not see anything like this.

1 Like

Just ran into this exact situation. FYI you can’t use “Jet Empty” liquid classes, use “Jet Part”:

1 Like

Hello,
Have you managed to determine how to select the pre-aspiration air volume?
I encountered the same issue and attempted to resolve it by enabling the labware properties mentioned. It pre-aspirates an undetermined volume of air before attaching the tip, yet the following problems remain:

  • I cannot find a way to select the pre-aspiration volume.
  • I am unable to initiate a dispense volume without first performing an aspirate step.
  • The pre-aspirated air is not being dispensed.
    (I also tried aspirating a zero volume with the blow-out tip dispense mode.)

I really appreciate any suggestion
Thank you!

The groups I know who used washers let the tips dry for a day or 2 before using.

Keep in mind the tip trays will eventually warp with repeated washes. Warping tip trays are no beuno.

Did the OP resolve this? I have some insight if needed. The answer is absolutely in the blowout tip / air gap settings in pipetting.

Personally, I’m not using washer, but I need to pre-aspirate air for another reason. Specifically: Pre-aspirate air (choosing the volume), Tip pick up , dispense air, tip eject.
Any suggestion would be appreciated

Oh you’re trying to aspirate air BEFORE tip pickup? That might be a challenge…may I ask why?

This might be something you’d need firmware commands for - I’m not sure if those exist where you could do something like…

Aspirate v vol @ z height with n tips
Get tips
Dispense v vol @ z height with n tips
Drop tips

You can trick the system to aspirate sans tip pickup by using error handling and empty tip wafers, but I’m not confident that the piston position would be preserved after dropping those fake tips to pick up real tips. Most likely the piston is zeroed after each tip eject.

Maybe something like this? If it works?

Get tips @ empty position
Error handling: ignore non-pickup
Aspirate v vol anywhere safe
Eject tips
Get real tips
Dispense v volume @ waste or something
1 Like

You can use ‘MlStarIsCoreHeadSpecialTipPickup’ property of the tip rack to pre-aspirate air that is then blown out AFTER picking up the tips. Setting ‘MlStarIsCoreHeadSpecialTipPickup’ = 1 will activate this property.

You can use ‘MlStarIsCoreHeadSpecialTipPickup’ = 2 to pre-aspirate and NOT blow out immediately (so that you can blowout into a waste or something with a dispense command) but I do not recommend using this because there are serious bugs associated with this property. Using “1” is safe though and does not require bug handling.

3 Likes

unfortunately it doesn’t work because, as you said, the piston is zeroed after each tip eject.

I realized there are serious bug with “2” !!

This property would ideally meet my requirements: to pre-aspirate without immediate blowout (possibly blowing out into waste or another position via a dispense command). Additionally, after picking up the tip, I want to avoid aspirating extra air, but only blow out the air that was pre-aspirated.

The challenges I encountered with this feature include:

  • There’s no option to select a pre-air volume (although this may not be a concern for me).

  • I’m compelled to perform a dummy aspiration (volume 0) AFTER picking up the tip, but I observed that the pre-aspirated air doesn’t blow out when using the dispense command.

Could you suggest a specific setting or liquid class that would allow for the blowout of pre-aspirated air?
I really appreciate your help

Yes absolutely is a challenge ! I need this for a particular application with phytips.

If I understand your use case, you can use ‘MlStarIsCoreHeadSpecialTipPickup’ = 2 to achieve your goal. This property will allow you to:

  1. Aspirate air
  2. Pickup a tip rack with the assigned property
  3. Follow the pickup with a ‘Blowout’ dispense step.

To circumvent the bug associated with this property you will need to follow these steps.

  1. Use the ‘Valueforkeysetforpropertiesoflabware’ step within the ‘LabwrAccess’ library to change the ‘MlStarIsCoreHeadSpecialTipPickup’ to 2 for the tip rack of interest.
  2. Perform your pickup and blowout.,
  3. Build in a property “reset” at the end of your method AND on abort. This reset involves again using ‘Valueforkeysetforpropertiesoflabware’ to set the ‘MlStarIsCoreHeadSpecialTipPickup’* to 1, performing a fake tip pickup/eject, then using the library to set ‘MlStarIsCoreHeadSpecialTipPickup’** to 0, then performing yet another fake tip pickup/blow.

You must go from 2 → 1 → 0 to achieve a proper property reset. If you do know perform the reset as outlined and complete/abort your method, ALL pickups will inherit a property of 2. This is not catastrophic but you will probably get “plunger out of bounds errors” and need to powercycle to reset the firmware.

Hope this helps.

2 Likes

@EricSindelar_Hamilton Is this on the VENUS bug fix list?

I’ve been using this MlStarIsCoreHeadSpecialTipPickup property and have lately run into an issue where the middle probes do not aspirate after tip pickup, outer edges of the 96 head work better but not great.

This was working fine a couple weeks ago but just started recently. We had service visit recently and they replaced spacer plates, which are supposed to be related but still this issues persists.

Curious if anyone has seen this before and has suggestions?