Cell straining on a liquid handler

Hi all, I’m trying to automate cell straining processes on liquid handlers for downstream flow cytometry (with access to Opentrons Flex and Hamilton Star). Are there any automation friendly 96-well based cell strainers? What is an appropriate pipetting speed for cell straining? Thanks in advance!

This looks to be similar material/pore size to the strainer caps for tubes:
Multiscreen® 96 well Plate, mesh nylon membrane pore size 40 μm, non-sterile | Sigma-Aldrich

They mention letting the cell suspension strain via gravity flow or gentle centrifugation

1 Like

We use this with great success:
CSCRIE

2 Likes

That looks amazing! Does it flow well by gravity or do you force the cell suspension through?

Pressure/Centrifuge is required to push the cells through the strainer.