I have encountered an issue with Dispense on the fly single step that I hope you can help me investigate.
The pipetting part of “Dispense on the fly” works flawlessly. However, after dispensing to a plate the system will stop and prompt the “risk for collision” dialog instead of performing an tip eject.
Trace says: [P8: Syntax, Parameter out of range.] and (Parameter out of range or missing parameter …)
(0x28 - 0x1 - 0x3)
I created a test method to investigate further but could not narrow it down further - it seems to be independent of the Dispense on the fly single step’s settings / advanced settings. I can provide logs for these 4 test case if anyone can provide an upload link.
Edit: I am pre-dispensing twice following aspiration, after dispensing there should be 1 post aliquot remaining (that should be thrown out with the tips).
I hope this is as trivial to fix as it usually are Thanks in advance
Can you upload the COM Trace? Specifically, the one named HxUsbComm20260609.trc. That’s the file that will give me the information to tell you what to look for.
@BrandonBare_Hamilton, I have tested without pre and post dispenses during early troubleshooting, but with same outcome. I just uploaded the com trace that also has the firmware command for debugging.