Labware properties list?

Hello! I had to make custom labware definitions for a carrier which incluses a .tml, 2 .rck, and 48 .ctr positions, but I am having issues getting the carrier to load with the autoload scanner.

I keep getting the error MlStarReadBarcodeOnPositions property is in the wrong format and I havent found a format that works yet. The Labware in question will only need to read one barcode on the rack, but defining the readable position hasnt worked so far. Is there a reference table of labware properties and their allowable formats available?

This came up recently for me too! The correct nomenclature that worked was β€œX=[Label]” where X is the position location of the carrier and β€œLabel” is the name of the position that should map to the Label column in the Template Sites of your .tml file. So for example, say you have a carrier with these labels

The MlStarReadBarcodeOnPositions property that corresponds to this carrier would be β€œ1=1_MIDI_HPTab;3=3_TipBC;4=4_TipBC;5=5_DWP_HPTab;”. Note that this example carrier has an empty position 2, so the property does not explicitly call out an assignment for X = 2.

Then when calling the Load Carrier step, in the advanced tab, you enter the Label of the labware that you would like to scan at that position:

Hopefully this helps!

1 Like

This is really awesome, I never would have figured that out!

1 Like