MCA96 Liquid Classes

Hey everyone,

I was hoping that FC 3.5 would have liquid classes for the MCA96 but it does not.

Since the MCA96 uses the FCA tips, I assume that I can start with the airFCA version of existing liquid classes. With that said, the microscripts for liquid classes are monsters so it for sure won’t be 1:1.

Any guidance on developing these liquid classes?

Best,

What type of liquid are you optimizing for?

From what I understood is that MCA96 liquid classes are the same as AirFCA, but there is more power in the MCA motor so some speeds could go higher

Additionally, the cLLD works with the first tip that touches a liquid. Other tips will follow accordingly.

Yeah it makes sense that it’s more like the AirFCA. There’s a MCA96 reference for the 10uL tip and one for 1mL tip but those tips vary in shape and form from the 50 or 200 tips. And I’ve been working with Fluent Control since version 1.3 so when it comes to pipetting, I am painfully aware that the devil is in the details.

Not having a reference available is a problem.

1 Like

I did some early work with the MCA96 head - 1 uL pipetting with 50 uL tips,
Got some pretty amazing accuracy and precision numbers
let me dig out the parameters and see how to share them

from the v3.5 liquid classes I saw, the provided classes are v basic, and only for v few tip types - a lot of our clients/Tecan customers have asked for my support in designing/optimizing the parameters for the tips/volumes deficient in the ones provided by default

happy to talk more details if you need something more bespoke

1 Like

Technically any liquid since the provided template LC’s are extremely limited.

They COULD but they don’t seem to even in the templates.

I’d be interested in seeing this as well to possibly translate your success to my Biomek instrument. 1µL pipetting is no joke.

1 Like

Happy to talk more Biomek offline if you’d like…I have done some optimizing for low volumes.

Curious but has anyone done a comparison of the detection values of the MCA96 vs the FCA?

I was able to receive a development set of liquid classes from Tecan. I’ll review them and share any key info.

With that said, I made my own smart LC’s from scratch to better utilize a wider array of functions. Cheers.

2 Likes

Initial impressions with the MCA96 are pretty good. I guess I took for granted what a game changer using a 1mL tip would be for a 96 well plate.

Partiap tip box mounting has had no issues, no accidental crashes so far but I’m making sure that the SBS holders and the tip boxes actually clip on.

In addition, it seems like the 96 channels are always “ON” so this create problems if you mount a partial tip box as if it were a full tip box and then use cLL D in a predominantly empty plate. With that said, if you partially mount tips and it knows where those tips are mounted, the detection values are pretty robust in a plate and you don’t get any wonky issues. I am able to achieve proper cLLD behavior which is very nice if you like to build smart custom LC’s with detection data.

2 Likes

I notice that for the MCA384 with 96 adapter the default max pipette speeds are much lower compared to the air FCA. For example, with 200uL tips the max speed is set to 124.5uL/s compared to 500uL/s for the FCA.

I would like to increase the robustness if my MCA mixing. Is there any concern with increasing this max speed up to the 200-500uL/s range? Is this limit just a guidance for precision pipetting or is there any hardware robustness concerns?

The MCA96 (which is a different head) has some default top speed’s as high as 600 ul/s for the dispense so you’re probably fine but the formula section of your microscript may hint at min & max values for speed.

Yes, in FC v3.5 the formulas set it at 125 uL/s as a max, but I believe permits me to overwrite those limits.

I was referencing this:

I am able to click into those min and max entries you’re showing in the formulas window and override the values. I just want to make sure that ability doesn’t let me break anything.

Tecan support claims "There is likely a risk of premature wear depending on how fast you set it, and of course accuracy. "

  • The maximum speed for the MCA384 arm is 125 μl/s, and the maximum acceleration/deceleration is 6250 μl/s². These values are referenced for the MCA384 head and are relevant for the 96 adapter as well. Exceeding these values may result in errors or suboptimal performance. EVOware is not able to check the validity of the parameters for the ramp settings of the MCA384, so manual adjustment is required if you build a custom liquid class.

  • For the Extended Volume Adapter (EVA, 4-to-1), the maximum speed is 132 μl/s and the maximum acceleration/deceleration is 6250 μl/s².

  • The default speed for the MCA384 is 95 μl/s, but you can increase up to the maximum values above for more intense mixing.

I’ve read through this a few times and am a new Tecan user. An automation script/method was shared with us for a Zymo extraction kit. But when trying to use some of the liquid classes the MCA doesn’t have any values or a microscript, this is really hindering us for some of the ethanol pieces.

1- How can edit the liquid classes? Even when duplicating, I can’t seem to change any values or enter any values
2- How do I make a microscript? Does it need to be as complicated as some of the others I’ve looked through? When I have tried to copy and paste from a similiar liquid class, I have received an error.
3- Ideally, I just want to pick up the desired liquid as is, no overage.

I can provide any other info I may be missing, thanks!

2 Likes

I’m curious if you ever shared these parameters? I’m trying to set up a 1.2uL multi dispense (wet contact) and it’s killing me! I’d be interested to know what you’re using for dispense speed/acceleration/deceleration and whether you’ve added a trailing air gap? I’m struggling to find the right combination of parameters to get such a small volume of liquid to consistently come out of the tip. I’m seeing a bunch of missed wells, followed by a double volume dispensed into the subsequent plate.

2 Likes