Positive Pressure SPE problem (Amplius Positive Pressure SPE Extractor)

Hi all,

We’re trying to automate the elution step on a positive-pressure SPE setup where we have 96-pressure channels on the head.

In the setup, we have 96 cartridges in an adapter, with a 1.3 mL deep-well plate underneath as the collection plate. I ramp the pressure from 100 mbar up to 600 mbar, hold at 600 mbar for some time, then lower it again.

With a low number of cartridges this mostly works fine, but with higher numbers (more than 32) I run into the following problem: foam forms at the cartridge outlets (the eluent is 50% acetonitrile), and this foam spreads across the deep-well plate, leading to non-uniform levels and cross-contamination. Notably, I don’t see this with water, it only happens with the 50% acetonitrile, which makes me suspect it’s related to the lower surface tension of the organic solvent.

Do you have any suggestions?

1 Like

What instrument?

It is Amplius Positive Pressure x well SPE Extractor

1 Like

I think you’re blowing for too long. Bubbles means air is passing through, lower the pressure and check the time for when the bulk of the volume comes through. Then shut it off.

1 Like