Request to purchase: Hamilton iSWAP calibration tool

I have yet to have any Hamilton sales rep reply to my requests to purchase this tool.

With 8+ STARs in production, iSWAP calibration drift due to crashes or labware mishandling force us to spend 40+ hours per month manually re-calibrating our iSWAP arms.

We know this can be done automatically with the iSWAP calibration tool, but there is no clear way to access this device, so instead we waste valuable engineer time aligning the arm by eye and re-teaching positions after every crash.

Does anyone know how we get one of these calibration tools?

Heya Ben,

That’s got to be frustrating.

Calibration tools tend to be quoted out by the service team, not the sales team. Generally sales refers us over to service for these things.

I’d send an email to RoboticsServicePartQuote@hamiltoncompany.com or RoboticService@hamiltoncompany.com

Also, I heave heard that the iSwap calibration, even with the tool, is a bear, so be warned.

Brad

2 Likes

Do you think it is possible to engineer better calibration tooling ourselves?

Fundamentally: is the iSWAP difficult to align, do to some inherent mechanical design decision?

to adjust iSWAP, you have to purchase 6 service tools and get the service documents. It is not that simple, the engineer should be trained. After training, the engineer can do these service works with these tools. For some OEM partners of Hamilton, like Roche/Ausbio/Bioxun, their engineers will repair/install/adjust STAR themselves. So it is possible

During adjustment, you have to connect a cable of service tool to the PCB of iSWAP when instrument is running, there is a high risk to damage the instrument. Also if the cable is not properly placed and fixed, it will block the rotation of iSWAP and this may misadjust iSWAP.



4 Likes

Thanks @Huajiang for the comprehensive instructions & @Brad for the right internal contact at Hamilton to get an immediate response with a quote.

This type of tacit knowledge could have saved us months had I known to ask you guys earlier - evidence of this forum pulling through heavily.

6 Likes

Realistically you only really need 3 and 5 from the items above. The “Adjust_Iswap_Above_Labware” uses the standard Star Tool. #1 is a square to verify that it is perpendicular to the deck. This could be 3D printed with enough accuracy to get you there.

I will stay away from 3D as it need to be precise.

1 Like

do you have more pictures showing the install process? i think i have everything right, i just don’t know exactly where to plug the sensor plug into:

wrong before, i think i have it installed correctly now! the signal cord is physically blocked from plugging in any other way