Ham
December 31, 2025, 8:03pm
1
Today, we had a Hamilton Star serviced for an unrelated reason (Core head crashed due to being thrown too far in x during a run, gantry rail was lubricated and adjustments were performed) and now I’m trying to perform testing to determine if the original issue has been resolved or not.
The other instruments with this same script are not having this issue, so I want to try to identify why this is happening to see if we can resolve it ourselves or if we need the FSE to intervene for this new issue.
Despite us having run this script in its entirety with this step as-is for years, it is suddenly “skipping” over the unload step for each carrier near the beginning of the script. As you can see in the log file, it is acknowledging that the unload step should be happening, but the Autoloader doesn’t respond to any of the requests and within 1 second the script continues on. Then, when it’s time for the carriers to be loaded back on the instrument (according to the order of the script), the Autoloader does move and try to line up to load a carrier back on deck.
The instrument can perform Weekly Maintenance without any issue, and the issue is replicable after a full power cycle and reinitialization.
Happy to provide any more information to try to get a better grasp on what could be causing this.
Thank you in advance and happy New Years Eve!
Hello @Ham
Would you be able to upload the communication trace from today? I need to review the responses from firmware to see what’s going on and why errors aren’t being thrown.
This file will be in the default LogFiles folder with the following prefix:
HxUsbComm(todays date).trc
Something that can happen during FSE interventions is that x, y, z coordinates change slightly. A classic problem that can occur then is that loading steps no longer work, as tracks are calculated via x coordinates and even if you’re just 0.1 mm in the neighboring track’s space, it will start looking for the carrier one track to the left or right. If your unloading steps have errorhandling that ignores missing carriers, it would look like you are skiping unloading the respective tracks. I would recommend writing a short test method that unloads a carrier on, say, track 10, but insert carriers on tracks 9, 10 and 11. Then you can check which carrier gets unloaded and see whether this is the problem you are facing. If it is, you don’t have to reteach everything, just call in the FSE again to repeat the deck adjustment macros more carefully.
Check the rear of the carrier and see if its damaged. Sometimes the metal piece is missing or damaged. Or see if the carrier isn’t push all the way into the deck it won’t be detected as present by the Hamilton. This causes the autoloader to think that there is no carrier loaded even if it is present on the deck.
Ham
January 2, 2026, 1:25pm
5
Yes! It’s not letting me attach the file itself to this thread since it’s not a photo, but below I’ve pasted the immediate information before, during, and after the time period of the attempted unloading per the screenshot in my initial post (14:37:52). Thank you.
< 14:37:32.479 8AF#8000#00: C0RQid0101
14:37:32.495 8AF#8000#00: C0RQid0101rq0000
< 14:37:32.496 8AF#8000#00: C0QBid0102
14:37:32.513 8AF#8000#00: C0QBid0102er00/00qb1
< 14:37:32.513 8AF#8000#00: C0RIid0103
14:37:32.529 8AF#8000#00: C0RIid0103er00/00si12/23/2022snE464
< 14:37:32.529 8AF#8000#00: C0QMid0104
14:37:32.567 8AF#8000#00: C0QMid0104er00/00ka000003xt54xa54xw13400xl05xr00xm03500xx11400ys090xu3540xv3700yu0060kl360kc0yx0060ke00000000xn00xo00ym6065kr0km360
< 14:37:32.568 8AF#8000#00: C0RMid0105
14:37:32.633 8AF#8000#00: C0RMid0105er00/00kb0Dkp08 C00000 X00000 P10000 P20000 P30000 P40000 P50000 P60000 P70000 P80000 I00000 H00000
< 14:37:32.634 8AF#8000#00: C0RFid0106
14:37:32.651 8AF#8000#00: C0RFid0106er00/00rf7.6S 25 2021_11_05 (GRU C0)
< 14:37:32.651 8AF#8000#00: P1RFid0107
14:37:32.662 8AF#8000#00: P1RFid0107rf4.0S j 2022-03-16
< 14:37:32.663 8AF#8000#00: P1RJid0108
14:37:32.674 8AF#8000#00: P1RJid0108jd2025-12-31js1
< 14:37:32.676 8AF#8000#00: P2RFid0109
14:37:32.687 8AF#8000#00: P2RFid0109rf4.0S j 2022-03-16
< 14:37:32.688 8AF#8000#00: P2RJid0110
14:37:32.700 8AF#8000#00: P2RJid0110jd2025-12-31js1
< 14:37:32.701 8AF#8000#00: P3RFid0111
14:37:32.713 8AF#8000#00: P3RFid0111rf4.0S j 2022-03-16
< 14:37:32.714 8AF#8000#00: P3RJid0112
14:37:32.727 8AF#8000#00: P3RJid0112jd2025-12-31js1
< 14:37:32.728 8AF#8000#00: P4RFid0113
14:37:32.739 8AF#8000#00: P4RFid0113rf4.0S j 2022-03-16
< 14:37:32.741 8AF#8000#00: P4RJid0114
14:37:32.753 8AF#8000#00: P4RJid0114jd2025-12-31js1
< 14:37:32.754 8AF#8000#00: P5RFid0115
14:37:32.765 8AF#8000#00: P5RFid0115rf4.0S j 2022-03-16
< 14:37:32.766 8AF#8000#00: P5RJid0116
14:37:32.779 8AF#8000#00: P5RJid0116jd2025-12-31js1
< 14:37:32.780 8AF#8000#00: P6RFid0117
14:37:32.792 8AF#8000#00: P6RFid0117rf4.0S j 2022-03-16
< 14:37:32.793 8AF#8000#00: P6RJid0118
14:37:32.805 8AF#8000#00: P6RJid0118jd2025-12-31js1
< 14:37:32.807 8AF#8000#00: P7RFid0119
14:37:32.818 8AF#8000#00: P7RFid0119rf4.0S j 2022-03-16
< 14:37:32.819 8AF#8000#00: P7RJid0120
14:37:32.831 8AF#8000#00: P7RJid0120jd2025-12-31js1
< 14:37:32.832 8AF#8000#00: P8RFid0121
14:37:32.843 8AF#8000#00: P8RFid0121rf4.0S j 2022-03-16
< 14:37:32.844 8AF#8000#00: P8RJid0122
14:37:32.856 8AF#8000#00: P8RJid0122jd2025-12-31js1
< 14:37:32.856 8AF#8000#00: C0QBid0123
14:37:32.874 8AF#8000#00: C0QBid0123er00/00qb1
< 14:37:32.875 8AF#8000#00: X0RFid0124
14:37:32.883 8AF#8000#00: X0RFid0124rf1.4S 2012-04-25
< 14:37:32.884 8AF#8000#00: X0RJid0125
14:37:32.893 8AF#8000#00: X0RJid0125jd2025-12-31js1
< 14:37:32.893 8AF#8000#00: I0RFid0126
14:37:32.903 8AF#8000#00: I0RFid0126rf3.4S f 2017-01-09
< 14:37:32.904 8AF#8000#00: I0RJid0127
14:37:32.914 8AF#8000#00: I0RJid0127jd2025-02-25js1
< 14:37:32.914 8AF#8000#00: H0RFid0128
14:37:32.923 8AF#8000#00: H0RFid0128rf5.0S j 2022-09-21 (H0 XE167)
< 14:37:32.924 8AF#8000#00: H0RJid0129
14:37:32.933 8AF#8000#00: H0RJid0129jd2025-12-31js1
< 14:37:32.935 8AF#8000#00: C0RMid0130
14:37:33.000 8AF#8000#00: C0RMid0130er00/00kb0Dkp08 C00000 X00000 P10000 P20000 P30000 P40000 P50000 P60000 P70000 P80000 I00000 H00000
< 14:37:33.013 8AF#8000#00: C0QMid0131
14:37:33.050 8AF#8000#00: C0QMid0131er00/00ka000003xt54xa54xw13400xl05xr00xm03500xx11400ys090xu3540xv3700yu0060kl360kc0yx0060ke00000000xn00xo00ym6065kr0km360
< 14:37:33.051 8AF#8000#00: H0QGid0132
14:37:33.059 8AF#8000#00: H0QGid0132qg2
< 14:37:33.060 8AF#8000#00: C0SRid0133
14:37:33.077 8AF#8000#00: C0SRid0133er00/00sr055
< 14:37:33.096 8AF#8000#00: C0RUid0134
14:37:33.129 8AF#8000#00: C0RUid0134er00/00ru00950 13400 30000 30000
< 14:37:40.288 8AF#8000#00: C0RVid0135vo00
14:37:40.349 8AF#8000#00: C0RVid0135er00/00vd2025-12-31 14:23vs1
< 14:37:40.350 8AF#8000#00: C0RVid0136vo01
14:37:40.411 8AF#8000#00: C0RVid0136er00/00vd2025-12-31 14:23vs1
< 14:37:40.413 8AF#8000#00: C0RSid0137
14:37:40.431 8AF#8000#00: C0RSid0137er00/00ts2025-12-31 14:23
< 14:37:40.433 8AF#8000#00: C0RIid0138
14:37:40.449 8AF#8000#00: C0RIid0138er00/00si12/23/2022snE464
< 14:37:40.451 8AF#8000#00: C0RMid0139
14:37:40.515 8AF#8000#00: C0RMid0139er00/00kb0Dkp08 C00000 X00000 P10000 P20000 P30000 P40000 P50000 P60000 P70000 P80000 I00000 H00000
< 14:37:40.521 8AF#8000#00: P1RVid0140
14:37:40.537 8AF#8000#00: P1RVid0140na0000067978nb0000057073nc0000077469nd0000109495
< 14:37:40.538 8AF#8000#00: P2RVid0141
14:37:40.553 8AF#8000#00: P2RVid0141na0000067771nb0000056908nc0000077159nd0000109306
< 14:37:40.554 8AF#8000#00: P3RVid0142
14:37:40.568 8AF#8000#00: P3RVid0142na0000067425nb0000056638nc0000076863nd0000109006
< 14:37:40.569 8AF#8000#00: P4RVid0143
14:37:40.589 8AF#8000#00: P4RVid0143na0000068847nb0000058027nc0000076534nd0000108537
< 14:37:40.591 8AF#8000#00: P5RVid0144
14:37:40.605 8AF#8000#00: P5RVid0144na0000068497nb0000057783nc0000076347nd0000108217
< 14:37:40.605 8AF#8000#00: P6RVid0145
14:37:40.621 8AF#8000#00: P6RVid0145na0000066853nb0000044193nc0000075930nd0000107921
< 14:37:40.622 8AF#8000#00: P7RVid0146
14:37:40.638 8AF#8000#00: P7RVid0146na0000080215nb0000065985nc0000075665nd0000107634
< 14:37:40.639 8AF#8000#00: P8RVid0147
14:37:40.654 8AF#8000#00: P8RVid0147na0000080979nb0000052229nc0000075562nd0000107425
< 14:37:40.655 8AF#8000#00: P1VWid0148
14:37:40.669 8AF#8000#00: P1VWid0148vw0 0 1 0
< 14:37:40.672 8AF#8000#00: P2VWid0149
14:37:40.685 8AF#8000#00: P2VWid0149vw0 0 1 0
< 14:37:40.687 8AF#8000#00: P3VWid0150
14:37:40.702 8AF#8000#00: P3VWid0150vw0 0 1 0
< 14:37:40.704 8AF#8000#00: P4VWid0151
14:37:40.718 8AF#8000#00: P4VWid0151vw0 0 1 0
< 14:37:40.720 8AF#8000#00: P5VWid0152
14:37:40.734 8AF#8000#00: P5VWid0152vw0 0 1 0
< 14:37:40.736 8AF#8000#00: P6VWid0153
14:37:40.749 8AF#8000#00: P6VWid0153vw0 0 1 0
< 14:37:40.751 8AF#8000#00: P7VWid0154
14:37:40.766 8AF#8000#00: P7VWid0154vw0 0 1 0
< 14:37:40.768 8AF#8000#00: P8VWid0155
14:37:40.781 8AF#8000#00: P8VWid0155vw0 0 1 0
< 14:37:40.808 8AF#8000#00: C0TTid0156tt44tf1tl0570tv01750tg3tu0
14:37:40.835 8AF#8000#00: C0TTid0156er00/00
< 14:37:40.844 8AF#8000#00: C0TTid0157tt33tf0tl0319tv00700tg6tu0
14:37:40.871 8AF#8000#00: C0TTid0157er00/00
< 14:37:40.883 8AF#8000#00: C0TTid0158tt22tf0tl0424tv00650tg2tu0
14:37:40.910 8AF#8000#00: C0TTid0158er00/00
< 14:37:40.924 8AF#8000#00: C0TTid0159tt11tf0tl0490tv00150tg2tu1
14:37:40.950 8AF#8000#00: C0TTid0159er00/00
< 14:37:40.966 8AF#8000#00: C0TTid0160tt03tf1tl0219tv00100tg1tu0
14:37:40.992 8AF#8000#00: C0TTid0160er00/00
< 14:37:41.007 8AF#8000#00: C0TTid0161tt45tf1tl0870tv03450tg3tu0
14:37:41.034 8AF#8000#00: C0TTid0161er00/00
< 14:37:41.047 8AF#8000#00: C0TTid0162tt34tf0tl0235tv00010tg2tu0
14:37:41.075 8AF#8000#00: C0TTid0162er00/00
< 14:37:41.089 8AF#8000#00: C0TTid0163tt23tf1tl0424tv00600tg2tu0
14:37:41.116 8AF#8000#00: C0TTid0163er00/00
< 14:37:41.131 8AF#8000#00: C0TTid0164tt12tf0tl0490tv03500tg2tu1
14:37:41.157 8AF#8000#00: C0TTid0164er00/00
< 14:37:41.167 8AF#8000#00: C0TTid0165tt04tf0tl0871tv12500tg3tu0
14:37:41.194 8AF#8000#00: C0TTid0165er00/00
< 14:37:41.205 8AF#8000#00: C0TTid0166tt46tf0tl0290tv17200tg3tu1
14:37:41.231 8AF#8000#00: C0TTid0166er00/00
< 14:37:41.243 8AF#8000#00: C0TTid0167tt35tf0tl0570tv03500tg3tu0
14:37:41.269 8AF#8000#00: C0TTid0167er00/00
< 14:37:41.284 8AF#8000#00: C0TTid0168tt24tf0tl0820tv01800tg3tu1
14:37:41.310 8AF#8000#00: C0TTid0168er00/00
< 14:37:41.324 8AF#8000#00: C0TTid0169tt13tf0tl0820tv12500tg3tu1
14:37:41.350 8AF#8000#00: C0TTid0169er00/00
< 14:37:41.367 8AF#8000#00: C0TTid0170tt05tf1tl0871tv10650tg3tu0
14:37:41.392 8AF#8000#00: C0TTid0170er00/00
< 14:37:41.406 8AF#8000#00: C0TTid0171tt47tf0tl0465tv00010tg2tu0
14:37:41.434 8AF#8000#00: C0TTid0171er00/00
< 14:37:41.448 8AF#8000#00: C0TTid0172tt36tf0tl0870tv03850tg3tu0
14:37:41.475 8AF#8000#00: C0TTid0172er00/00
< 14:37:41.488 8AF#8000#00: C0TTid0173tt25tf0tl1080tv54200tg5tu0
14:37:41.515 8AF#8000#00: C0TTid0173er00/00
< 14:37:41.529 8AF#8000#00: C0TTid0174tt14tf0tl0220tv00010tg0tu0
14:37:41.554 8AF#8000#00: C0TTid0174er00/00
< 14:37:41.571 8AF#8000#00: C0TTid0175tt06tf0tl0519tv00700tg2tu1
14:37:41.596 8AF#8000#00: C0TTid0175er00/00
< 14:37:41.611 8AF#8000#00: C0TTid0176tt48tf0tl0829tv08000tg3tu1
14:37:41.640 8AF#8000#00: C0TTid0176er00/00
< 14:37:41.653 8AF#8000#00: C0TTid0177tt37tf0tl0319tv00700tg6tu0
14:37:41.679 8AF#8000#00: C0TTid0177er00/00
< 14:37:41.694 8AF#8000#00: C0TTid0178tt26tf0tl1080tv00010tg5tu0
14:37:41.720 8AF#8000#00: C0TTid0178er00/00
< 14:37:41.735 8AF#8000#00: C0TTid0179tt15tf0tl0256tv00350tg4tu0
14:37:41.761 8AF#8000#00: C0TTid0179er00/00
< 14:37:41.775 8AF#8000#00: C0TTid0180tt07tf0tl0519tv03500tg2tu1
14:37:41.802 8AF#8000#00: C0TTid0180er00/00
< 14:37:41.817 8AF#8000#00: C0TTid0181tt49tf0tl0820tv03020tg3tu1
14:37:41.843 8AF#8000#00: C0TTid0181er00/00
< 14:37:41.857 8AF#8000#00: C0TTid0182tt38tf0tl0490tv00650tg2tu1
14:37:41.883 8AF#8000#00: C0TTid0182er00/00
< 14:37:41.899 8AF#8000#00: C0TTid0183tt27tf0tl0190tv00010tg0tu0
14:37:41.925 8AF#8000#00: C0TTid0183er00/00
< 14:37:41.938 8AF#8000#00: C0TTid0184tt16tf0tl0251tv00350tg0tu0
14:37:41.965 8AF#8000#00: C0TTid0184er00/00
< 14:37:41.980 8AF#8000#00: C0TTid0185tt08tf0tl0871tv12500tg3tu1
14:37:42.006 8AF#8000#00: C0TTid0185er00/00
< 14:37:42.020 8AF#8000#00: C0TTid0186tt39tf0tl0820tv01200tg3tu1
14:37:42.046 8AF#8000#00: C0TTid0186er00/00
< 14:37:42.059 8AF#8000#00: C0TTid0187tt28tf0tl0626tv04000tg6tu0
14:37:42.087 8AF#8000#00: C0TTid0187er00/00
< 14:37:42.096 8AF#8000#00: C0TTid0188tt17tf0tl0251tv00350tg0tu0
14:37:42.123 8AF#8000#00: C0TTid0188er00/00
< 14:37:42.132 8AF#8000#00: C0TTid0189tt29tf1tl1080tv43670tg5tu0
14:37:42.157 8AF#8000#00: C0TTid0189er00/00
< 14:37:42.168 8AF#8000#00: C0TTid0190tt18tf1tl0210tv12500tg0tu1
14:37:42.195 8AF#8000#00: C0TTid0190er00/00
< 14:37:42.207 8AF#8000#00: C0TTid0191tt19tf0tl0534tv00670tg2tu0
14:37:42.234 8AF#8000#00: C0TTid0191er00/00
< 14:37:42.248 8AF#8000#00: C0TTid0192tt50tf0tl0820tv08000tg3tu1
14:37:42.274 8AF#8000#00: C0TTid0192er00/00
< 14:37:42.288 8AF#8000#00: C0TTid0193tt51tf0tl0111tv00010tg0tu0
14:37:42.315 8AF#8000#00: C0TTid0193er00/00
< 14:37:42.330 8AF#8000#00: C0TTid0194tt40tf0tl0820tv02000tg3tu1
14:37:42.355 8AF#8000#00: C0TTid0194er00/00
< 14:37:42.368 8AF#8000#00: C0TTid0195tt52tf0tl0430tv00010tg0tu0
14:37:42.394 8AF#8000#00: C0TTid0195er00/00
< 14:37:42.408 8AF#8000#00: C0TTid0196tt41tf0tl0820tv03460tg3tu1
14:37:42.434 8AF#8000#00: C0TTid0196er00/00
< 14:37:42.450 8AF#8000#00: C0TTid0197tt30tf0tl0519tv04000tg2tu0
14:37:42.476 8AF#8000#00: C0TTid0197er00/00
< 14:37:42.489 8AF#8000#00: C0TTid0198tt00tf0tl0519tv04000tg2tu0
14:37:42.515 8AF#8000#00: C0TTid0198er00/00
< 14:37:42.526 8AF#8000#00: C0TTid0199tt53tf0tl0485tv35000tg0tu0
14:37:42.552 8AF#8000#00: C0TTid0199er00/00
< 14:37:42.564 8AF#8000#00: C0TTid0200tt42tf0tl0490tv03550tg2tu1
14:37:42.590 8AF#8000#00: C0TTid0200er00/00
< 14:37:42.601 8AF#8000#00: C0TTid0201tt31tf0tl0422tv00650tg2tu0
14:37:42.628 8AF#8000#00: C0TTid0201er00/00
< 14:37:42.641 8AF#8000#00: C0TTid0202tt20tf0tl0343tv00650tg4tu0
14:37:42.668 8AF#8000#00: C0TTid0202er00/00
< 14:37:42.686 8AF#8000#00: C0TTid0203tt01tf1tl0519tv03600tg2tu0
14:37:42.712 8AF#8000#00: C0TTid0203er00/00
< 14:37:42.724 8AF#8000#00: C0TTid0204tt43tf0tl0820tv11350tg3tu1
14:37:42.751 8AF#8000#00: C0TTid0204er00/00
< 14:37:42.765 8AF#8000#00: C0TTid0205tt32tf0tl0219tv00150tg1tu0
14:37:42.795 8AF#8000#00: C0TTid0205er00/00
< 14:37:42.807 8AF#8000#00: C0TTid0206tt21tf0tl0442tv00610tg6tu1
14:37:42.834 8AF#8000#00: C0TTid0206er00/00
< 14:37:42.847 8AF#8000#00: C0TTid0207tt10tf0tl0519tv04000tg2tu0
14:37:42.873 8AF#8000#00: C0TTid0207er00/00
< 14:37:42.888 8AF#8000#00: C0TTid0208tt02tf0tl0219tv00150tg1tu0
14:37:42.914 8AF#8000#00: C0TTid0208er00/00
< 14:37:42.915 8AF#8000#00: C0DRid0209
14:37:42.937 8AF#8000#00: C0DRid0209er00/00
< 14:37:42.939 8AF#8000#00: C0CUid0210cu0
14:37:42.959 8AF#8000#00: C0CUid0210er00/00
< 14:37:42.962 8AF#8000#00: C0CEid0211
14:37:42.982 8AF#8000#00: C0CEid0211er00/00
< 14:37:42.984 8AF#8000#00: C0COid0212
14:37:43.196 8AF#8000#00: C0COid0212er00/00
< 14:37:43.198 8AF#8000#00: C0AWid0213
14:37:43.550 8AF#8000#00: C0AWid0213er00/00
< 14:37:43.554 8AF#8000#00: PXAFid0214af0
14:37:43.687 8AF#8000#00: PXAFid0214er00
< 14:37:43.690 8AF#8000#00: C0QSid0215
14:37:43.734 8AF#8000#00: C0QSid0215er00/00qs0 0 0 0 0 0 0 0
< 14:37:43.735 8AF#8000#00: C0RJid0216
14:37:43.781 8AF#8000#00: C0RJid0216er00/00tq1 1 1 1 1 1 1 1
< 14:37:43.786 8AF#8000#00: C0AWid0217
14:37:44.135 8AF#8000#00: C0AWid0217er00/00
< 14:37:44.140 8AF#8000#00: C0CPid0218cl00000000000000cb00000000000000
14:37:44.176 8AF#8000#00: C0CPid0218er00/00
< 14:37:44.179 8AF#8000#00: C0CBid0219bt7Fmq00
14:37:44.211 8AF#8000#00: C0CBid0219er00/00
< 14:37:44.360 8AF#8000#00: C0QWid0220
14:37:44.412 8AF#8000#00: C0QWid0220er00/00qw1
< 14:37:44.422 8AF#8000#00: C0RTid0221
14:37:44.465 8AF#8000#00: C0RTid0221er00/00rt0 0 0 0 0 0 0 0
< 14:37:44.467 8AF#8000#00: C0QHid0222
14:37:44.487 8AF#8000#00: C0QHid0222er00/00qh0
< 14:37:44.511 8AF#8000#00: C0CTid0223cp54
14:37:45.285 8AF#8000#00: C0CTid0223er00/00ct0
< 14:37:45.320 8AF#8002#00: T0QUid0001
14:37:45.326 8AF#8002#00: T0QUid0001au0 0 0 0 0 0 0 0 0 0
< 14:37:45.328 8AF#8002#00: T0QWid0002
14:37:45.330 8AF#8002#00: T0QWid0002qw1
< 14:37:45.332 8AF#8002#00: T0RFid0003
14:37:45.335 8AF#8002#00: T0RFid0003rf3.2S e 2015-08-17 (Heater Shaker)
< 14:37:45.338 8AF#8002#00: T1QUid0004
14:37:45.353 8AF#8002#00: T1QUid0004au0 0 0 0 0 0 0 0 0 0
< 14:37:45.355 8AF#8002#00: T1QWid0005
14:37:45.368 8AF#8002#00: T1QWid0005qw1
< 14:37:45.369 8AF#8002#00: T1RFid0006
14:37:45.384 8AF#8002#00: T1RFid0006rf3.2S e 2015-08-17 (Heater Shaker)
< 14:37:45.388 8AF#8002#00: T2QUid0007
14:37:45.404 8AF#8002#00: T2QUid0007au0 0 0 0 0 0 0 0 0 0
< 14:37:45.405 8AF#8002#00: T2QWid0008
14:37:45.419 8AF#8002#00: T2QWid0008qw1
< 14:37:45.420 8AF#8002#00: T2RFid0009
14:37:45.434 8AF#8002#00: T2RFid0009rf3.2S e 2015-08-17 (Heater Shaker)
< 14:37:45.438 8AF#8002#00: T3QUid0010
14:37:45.453 8AF#8002#00: T3QUid0010au0 0 0 0 0 0 0 0 0 0
< 14:37:45.454 8AF#8002#00: T3QWid0011
14:37:45.468 8AF#8002#00: T3QWid0011qw1
< 14:37:45.470 8AF#8002#00: T3RFid0012
14:37:45.484 8AF#8002#00: T3RFid0012rf3.2S e 2015-08-17 (Heater Shaker)
< 14:37:46.825 8AF#8002#00: T0TOid0013
14:37:46.828 8AF#8002#00: T0TOid0013er00
< 14:37:46.833 8AF#8002#00: T0TAid0014ta0220tb1800tc0020td0060
14:37:46.842 8AF#8002#00: T0TAid0014er00
< 14:37:46.846 8AF#8002#00: T1TOid0015
14:37:46.852 8AF#8002#00: T1TOid0015er00
< 14:37:46.858 8AF#8002#00: T1TAid0016ta0220tb1800tc0020td0060
14:37:46.878 8AF#8002#00: T1TAid0016er00
< 14:37:46.882 8AF#8002#00: T2TOid0017
14:37:46.893 8AF#8002#00: T2TOid0017er00
< 14:37:46.898 8AF#8002#00: T2TAid0018ta0220tb1800tc0020td0060
14:37:46.912 8AF#8002#00: T2TAid0018er00
< 14:37:46.916 8AF#8002#00: T3TOid0019
14:37:46.928 8AF#8002#00: T3TOid0019er00
< 14:37:46.933 8AF#8002#00: T3TAid0020ta0220tb1800tc0020td0060
14:37:46.949 8AF#8002#00: T3TAid0020er00
< 14:37:52.178 8AF#8000#00: C0CPid0224cl00000000000007cb00000000000007
14:37:52.215 8AF#8000#00: C0CPid0224er00/00
< 14:37:52.216 8AF#8000#00: C0RCid0225
14:37:52.467 8AF#8000#00: C0RCid0225er00/00cd10400002082082ce00000000000010400002082082
< 14:37:52.469 8AF#8000#00: C0CPid0226cl00000000000000cb00000000000000
14:37:52.506 8AF#8000#00: C0CPid0226er00/00
< 14:37:52.529 8AF#8000#00: C0CPid0227cl000000000001F8cb000000000001F8
14:37:52.565 8AF#8000#00: C0CPid0227er00/00
< 14:37:52.567 8AF#8000#00: C0RCid0228
14:37:52.883 8AF#8000#00: C0RCid0228er00/00cd10400002082082ce00000000000010400002082082
< 14:37:52.888 8AF#8000#00: C0CPid0229cl00000000000000cb00000000000000
14:37:52.923 8AF#8000#00: C0CPid0229er00/00
< 14:37:52.938 8AF#8000#00: C0CPid0230cl00000000007E00cb00000000007E00
14:37:52.975 8AF#8000#00: C0CPid0230er00/00
< 14:37:52.977 8AF#8000#00: C0RCid0231
14:37:53.191 8AF#8000#00: C0RCid0231er00/00cd10400002082082ce00000000000010400002082082
< 14:37:53.193 8AF#8000#00: C0CPid0232cl00000000000000cb00000000000000
14:37:53.229 8AF#8000#00: C0CPid0232er00/00
< 14:37:53.243 8AF#8000#00: C0CPid0233cl000000001F8000cb000000001F8000
14:37:53.280 8AF#8000#00: C0CPid0233er00/00
< 14:37:53.281 8AF#8000#00: C0RCid0234
14:37:53.596 8AF#8000#00: C0RCid0234er00/00cd10400002082082ce00000000000010400002082082
< 14:37:53.602 8AF#8000#00: C0CPid0235cl00000000000000cb00000000000000
14:37:53.639 8AF#8000#00: C0CPid0235er00/00
< 14:37:53.651 8AF#8000#00: C0CPid0236cl00000007E00000cb00000007E00000
14:37:53.688 8AF#8000#00: C0CPid0236er00/00
< 14:37:53.689 8AF#8000#00: C0RCid0237
14:37:53.905 8AF#8000#00: C0RCid0237er00/00cd10400002082082ce00000000000010400002082082
< 14:37:53.912 8AF#8000#00: C0CPid0238cl00000000000000cb00000000000000
14:37:53.947 8AF#8000#00: C0CPid0238er00/00
< 14:37:53.960 8AF#8000#00: C0HOid0239
14:37:53.975 8AF#8000#00: C0HOid0239er00/00
< 14:37:58.660 8AF#8000#00: C0HDid0240
< 14:37:58.660 8AF#8000#00: C0RTid0241
14:37:58.700 8AF#8000#00: C0HDid0240er00/00
14:37:58.749 8AF#8000#00: C0RTid0241er00/00rt0 0 0 0 0 0 0 0
< 14:37:58.751 8AF#8000#00: C0QHid0242
14:37:58.771 8AF#8000#00: C0QHid0242er00/00qh0
< 14:37:58.779 8AF#8000#00: C0QHid0243
14:37:58.799 8AF#8000#00: C0QHid0243er00/00qh0
< 14:37:59.080 8AF#8002#00: T0SCid0021
14:37:59.843 8AF#8002#00: T0SCid0021er00
< 14:37:59.849 8AF#8002#00: T0LPid0022lp0
14:37:59.898 8AF#8002#00: T0LPid0022er00
< 14:37:59.904 8AF#8002#00: T0TOid0023
14:37:59.906 8AF#8002#00: T0TOid0023er00
< 14:37:59.912 8AF#8002#00: T1SCid0024
14:38:00.692 8AF#8002#00: T1SCid0024er00
Ham
January 2, 2026, 1:27pm
6
Thank you for the info, that’s good to keep in mind. In our case, based on the log file it seems like the script is auto ‘continuing’ through whatever error it’s possibly encountering in the background, however the unload step’s error handling is all default settings, none of which are continue. Execution error handling is set to repeat by default but it’s not attempting to repeat at all.
Ham
January 2, 2026, 1:28pm
7
Thank you, we checked that first but that is good information to include for others who may encounter unloading issues!
Roooo
January 4, 2026, 7:15am
8
I don’t see firmware errors during the unloading timeframe.
The carrier sensors are checked 5 times and always return:
I think you’re supposed to convert the return value from hex to binary to figure out which tracks it detects carriers on which yields
00010000010000000000000000000010000010000010000010000010
corresponding to tracks 2, 8, 14, 20, 26, 47, 53
There are also multiple commands to turn on the cover lights
converting those from hex to binary yields:
111
111111000
111111000000000
111111000000000000000
111111000000000000000000000
Tracks 1-3, 4-9, 10-15, 16-21, 22-27
Not sure what the expected track numbers your carriers are in, but as a previous poster mentioned, worth trying with your carriers shifted over a track. Hopefully a Hamilton forum member will correct if I’ve misinterpreted how to read these firmware responses.
Ham
January 5, 2026, 1:49pm
9
Thank you for the insight! I did try shifting over a track but unfortunately the same behavior persisted. The FSE will be here today, so hopefully the issue will be resolved.