Solvent dispensing

Anyone used 406’s syringe pumps to dispense solvents (e.g. Methanol or Acetonitrile)? Did you just repace the silicone tubing and check-valves? Anything in the syringe assembly?

Thanks,
Rob

I actually had some experience in this, we used liquid based Tecans to do dispense acetonitrile mixtures. Since the reagent didn’t make contact with the syringe and only the tubing and tips (I hope :slight_smile: ), we only did the usual maintenance on the Tecan. Our check values, and tubing still had to be replaced due to wear and tear.