** SOS ** cLLD Trouble Hamilton STAR (Urgent)
@EricSindelar_Hamilton @BrandonBare_Hamilton @NickHealy_Hamilton
Hamilton experts, I need help on this issue ASAP in hopes that’s possible (very important).
1000 µL (non-sterile CO-RE, without filter) tips are transferred to tip adapter in order to utilize H12 cLLD for the Head. Pipette step set High cLLD sensitivity (0 Aspiration mode).
Tips only needed to make one transfer (800 uL to 9 wells), row container had 17 mL.
This method/step ran on multiple different units, numerous times where this did not happen but randomly tips were unable to detect any liquid & kept reaching down to the point where it was pressing aggressively down on the container & carrier to the point where the tips were bent.
Container Info:
Axygen High Profile 8-Channel Trough Reservoir: Capacity 240 mL (30 mL/trough or row), Dead volume 2.5-3 mL.
Labware:
PLT_CAR_L5TAB_DWP, precision tab plate carrier
2 mL Nunc 96 Deep Well Plate
Row had plenty of volume needed but either way, why did the cLLD not work suddenly? Why would it keep going down?
I’ve tested it before with less volume where it worked yet confusing.
I have to now explain/justify this & prove that it won’t happen again. The only solution I have is to add more volume than originally required (20 mL) but still don’t understand what happened nor why, please help!
I’ve contacted your local support team to assist with your investigation. Based on the information provided, the root cause of the issue is not clear to me. We’ll work with them to help resolve it.
For urgent issues, please use the Hamilton Service Hotline: 1-800-527-5269 or roboticservice@hamiltoncompany.com.
Definitely let us know what the problem was. I saw something similar in two different cases:
  1. Tips were loaded in the incorrect spot, the MPH head cannot decern the difference between tip types like the 1mL channels can and this can lead to the Y direction being off.
  2. Labware definitions were incorrect — Especially for reagent reservoirs I’ve had some difficulty with some definitions not being exactly right. if you are going to bottom aspirate and the definition is off by a few mm this can happen.
You can test the latter by “re-teaching” the labware to check the labware bottom height is correct using the metal teaching needles.
I hope you found the resolution to your issue and you’re in good hands with the service team!
Do note that cLLD is only detected on A1/B2 and G11/H12 so these positions need to have a tip.
If one of these positions to not hold a tip, try to use a firmware command to disable the cLLD for these positions.
Also make sure your teaching is correct and that - obviously to say - you are selecting the correct labware to aspirate from.
By default I believe its only A1 & H12, you have to firmware command to turn on all 4 sensing channels.