Tecan Fluent error - parameter out of range
I am getting an error when executing a gwl in simulation. It does not like some height parameter. Anyone know where this height parameter is coming from and how to adjust it?
This is line 21 of my Script.
for simulation, i find it best to not use liquid detection - the microscript variable for liquid level isn’t set and the “height” error appears
the line 8 refers to the liquid class microscript you are using
Thanks for the fast reply! That makes sense. Going forward I’ll disable LLD in Sim.
In simulation mode, liquid level detection is triggered at Z-Start. This error message occurs in simulation when Z-Start is defined outside of the compartment definition. If the compartment is not defined correctly, this error can be observed in a real run as well. To verify that Z-Start is within the compartment, subtract Z-Bottom from Z-Start, and compare that to the total height of the compartment. The values of interested are circled below:
Wow that worked. The difference between Z Start and Bottom was just a few hundredths larger than the Compartment. I updated it so that it was equal to and it works without the error now. Great to know!
The Fluent software is insanely powerful but if you don’t get those little details with dimensions or definitions, you’ll get all sorts of wonky behavior. STANDARDIZE!
It also helps to audit your definitions.
I’ve just encountered this error. But luckily I had just read this thread so thanks @EricScheitz_Tecan for your answer and @tgraham for having already raised this. Much appreciated!
@EricScheitz_Tecan and @tgraham count another person thankful for running across this thread 1.5 years later!
Eric’s fix did the trick; reducing the z start slightly lower so the Z-Start minus Z-bottom value matched the compartment size did the trick.
I had to do it twice though, as the value displayed on the graphic rounds up from the total parts of your compartment. Leaving this here to help others in the future as well!
Just adding this error description in text so I can find it again if I, or anyone else need it:
Exception occurred during execution of pipetting
Specified argument was out of range of valid values
Parameter name: height