In simulation mode, liquid level detection is triggered at Z-Start. This error message occurs in simulation when Z-Start is defined outside of the compartment definition. If the compartment is not defined correctly, this error can be observed in a real run as well. To verify that Z-Start is within the compartment, subtract Z-Bottom from Z-Start, and compare that to the total height of the compartment. The values of interested are circled below: