Tips on adjusting X position of Te-Chrom Shuttle?

I’ve got an older EVO system running a Te-Chrom shuttle and the “MoveToStart” command puts the plate ~2mm off from directly beneath the column drip path.

I can’t seem to find how to adjust that alignment in the software.

EVOware 2.7.5.0

Thanks for any help.

go to configure > find driver for te-chrom shuttle > settings >

here you can change the x coordinate for the start & transfer positions

if you want to move the shuttle for verification, use firmware command Q1PAX1550 via the “Direct Command” (assuming device address is Q1)

this will move shuttle to the “middle” - adjust 1550 by 10 for every millimeter
DM me if you have questions

OK thanks!

The direct command works and I was able to align the plate to where it needed to be, but I can’t adjust the X coordinate associated with the command.

I might be able to get around this by dropping a direct command after the plate is moved to the first column.