Vantage loading tray fails to open (Venus 6.0.2)

We’ve had this “C0: Process busy” error opening the loading tray at the end of the method occur multiple times on one of our systems. I can easily add an explicit Lock/Unlock step before cleanup if this is the best course of action.

2026-01-06 13:42:50.865 SYSTEM : Execute method - complete;
2026-01-06 13:42:50.865 SYSTEM : End method - start;
2026-01-06 13:42:50.987 Microlab® STAR / VANTAGE : End method command - start;
2026-01-06 13:42:50.998 Microlab® STAR / VANTAGE : End method command - progress; TADM curve upload is in progress…
2026-01-06 13:42:59.371 Microlab® STAR / VANTAGE : Clean up instrument - progress; The instrument now tries to clean up its devices and superfluous material. Some of these items might be mentioned even if they are not configured or used.
2026-01-06 13:42:59.781 Microlab® STAR / VANTAGE : Clean up instrument - progress; Move up 1000ul channels
2026-01-06 13:42:59.963 Microlab® STAR / VANTAGE : Clean up instrument - progress; Move up CO-RE 96 Head
2026-01-06 13:43:00.176 Microlab® STAR / VANTAGE : Clean up instrument - progress; Move down auto load Y drive
2026-01-06 13:43:00.192 Microlab® STAR / VANTAGE : Clean up instrument - progress; Spreading channels
2026-01-06 13:43:01.952 Microlab® STAR / VANTAGE : Clean up instrument - progress; Switch off loading lights
2026-01-06 13:43:01.969 Microlab® STAR / VANTAGE : Clean up instrument - progress; Switch status light to idle.
2026-01-06 13:43:01.998 Microlab® STAR / VANTAGE : Lock/Unlock Front Cover (Single Step) - progress; Start error handling in walkaway mode (no dialog).
2026-01-06 13:43:01.998 Microlab® STAR / VANTAGE : Main - error; An error occurred while running Vector. The error description is: Step canceled. User-defined error handling will be used. (C0: Process busy.) (0x28 - 0x1 - 0x80a)
2026-01-06 13:43:01.998 Microlab® STAR / VANTAGE : Lock/Unlock Front Cover (Single Step) - progress; Error automatically recovered depending of custom error configuration (Walk-Away).
2026-01-06 13:43:03.686 Microlab® STAR / VANTAGE : End method command - complete;
2026-01-06 13:43:03.781 SYSTEM : End method - complete;
2026-01-06 13:43:03.851 SYSTEM : File checksum - written; $$author=Hamilton$$valid=0$$time=2026-01-06 13:43$$checksum=d884f3f9$$length=140$$

Hi @pthatcher,

This was a known bug in Venus 6 versions prior to Venus 6.2.1. In some cases the Vantage cleanup steps were not executing with correct timing resulting in the process busy error. Are your other Vantage systems also on Venus 6.0.2? Do you have the ability to upgrade to a newer version of Venus? Otherwise executing an unlock step prior to cleanup is the best route to go.

Matt

2 Likes

Hi @MatthewSmith_Hamilton, thanks for the guidance! We can easily drop a discrete step in before cleanup. I’ll take this back to the team to decide if we can just upgrade instead. Thanks!

1 Like