Venus 6 Smart Step Tip Loading

Not sure if this is user error or something else.

I am trying to load a carrier of tips using the STAR Smart Step. I am putting the relevant information into the box that pops up (i.e. Sequence, Checking the box for the Tip Counter, and putting the appropriate tip counter). With the program written as is, I run the method and it gets to the tip sequence loading step and immediately aborts with the following message, VENUS execution ended abruptly. No other hints as to what has happened. Has anyone seen this type of error and how to fix it?

Can you upload the trace file and method pkg so we can help troubleshoot further?

You can use this link.

Password is Hamilton23

1 Like

I uploaded the .pkg files and the .trc file for the issue I am having

Hi @Lost_in_Automation ,

This appears to be an interaction with 2D mode, the tip editor and our new Run Control.

I was only able to reproduce your issue by switching to 2D mode, however reverting back to 3D mode resulted in the steps performing as expected.

Verify that your Deck View mode is set to 3D and not 2D.

We have let product management know of this issue.

3 Likes

Thank you @BrandonBare_Hamilton for looking into this. When I set the Deck View Mode to 3D and go into method editor I get the following error:
Venus Error

I have uninstalled and reinstalled Venus 6 and made sure all Windows updates are up to date.

Hi @Lost_in_Automation ,

It looks like Direct X is failing to register. If the VENUS 6 reinstallation did not resolve the issue, I would recommend trying to manually install Direct X. It can be found within the VENUS installer subfolders.

image

Direct X is what VENUS uses to render the 3D deck view. Without it, you will continue to see the error you are seeing.

3 Likes

That was it! Thank you!

3 Likes

Did the Hamilton Software team fix this bug?