Have you ever had any experience working with 001-12782 (LP-0200) using only CO-RE 96? What tips have you used, liquid classes, carriers anything…?
We did work only with 50μl Standard Tips, because they match the wells when the head goes down into it… This tips are shorter than 300μl Standard Tips and for our needs they’d be more helpful…so we tried 300μl Standard Tips but had no result, because when the tips touched with wells they didn’t match and there was a sound of creaking… I hope you got it
I’m just wondering if you’re working with this kinda stuff like this…
If you’all want I can do the video too.
Tips 50 look pretty well, Tips 300 are creaking, Tips 300 slim are not recommended to use in 384 well plate at all (It says in Operator’s Manual page 42)
Oh, now I get where the creaking sound is coming from. It’s some of the larger tips scratching the inner wall of the wells. I initally thought it happened when the tips reached the bottom of the wells.
Hm, this is tricky, since the tips on the MPH are never aligned perfectly and I guess this is the reason why the slim tips are not recommended in combination with 384 well plates. Sorry for not being more helpful.
From time to time we work with plates such as Matrix 1.4 or Micronic 1.4 and you can’t reach low volumes using Tips 50, but Tips 300 perfect match the bottom
I just did this. The First step is that the tips reach “Touch-off at bottom height” which is for example 2 mm and the Second is the “Side touch”… I tried with 50tips and again it’s perfect… with others I’m scared)