Hi @Pascal,
Thanks for the response! It’s dispensing the same volume that it aspirates and the trough still has a good amount of liquid in it (+100 mLs). Seems like someone else had this issue recently as well and posted about it here.
I think it has something to do with the plunger position as you suggested because of the disappearance of the error when swapping to pLLD. Maybe there is some positional drift going on, and because I’m transferring 1000ul the plunger would go above it’s capacity and the Ham stops itself. It’d be great if there was an option to reinitialize the plungers mid run to recover from this.
Thanks,
Connor